Mobility calculation formulas
Apparent mobility of analyte A, mapp,A is calculated according to equation (1):
Leff - effective capillary length (from injection end to the detector position)
Ltot - total capillary length
Usep - applied separation voltage
tmig,A - migration time of the analyte A (in Gaussian peaks = migration time of peak apex).
Effective electrophoretic mobility, meff,A, is defined by eq. (2):
mEOF is mobility of electroosmotic flow (EOF) calculated according to eq. (3):
tmig,EOF - migration time of the electroneutral compound (marker of EOF)
Substitution of equations (1) and (3) into equation (2) gives the following equation for effective electrophoretic mobility of analyte A, meff,A:
Alternatively, mobility of EOF, meof, can be determined from the apparent and effective mobilities of a reference analyte R, mapp,R and meff,R, respectively.
mapp,R is calculated according to eq. (1), where tmig,A is replaced for migration time of reference analyte R, tmig,R:
Substitution of equations (6) into equation (5) gives the following equation for the electroosmotic flow mobility, mEOF:
Effective mobility of reference compound R, meff,R, has to be known from the previous measurement in the same background electrolyte and at the same temperature, at which the apparent mobility of analyte A is measured.
Effective electrophoretic mobility, meff,A, is then calculated according to eq. (2), where mapp,A is obtained from eq. (1) and the electroosmotic flow mobility, mEOF from eq. (7).
For the special case, when reference analyte R is electroneutral compound (EOF marker), i.e. meff,R = 0 and tmig,R = tmig,EOF, the eq. (7) becomes identical to eq. (3) and effective mobility of analyte A can be calculated according to eq. (4).